- Discussion:
- sliding screw fixation allows for improved fracture fixation w/ potential for
continuous impaction at frx site while lessening chance of nail penetration
thru or cutting out of the femoral head;
- important that the barrel must cross the fracture site and that sufficient
slide must remain in the nail after insertion;
- permit deeper insertion of screw w/o fear of later penetration of joint;
- permit controlled collapse of frx site w/o penetration of femoral head;
- improves wt bearing capacity of implant by reduction of moment arm;
- telescoping nails maximize bony contact & hence frx stability, thereby
decrease implant failure;
- Indications:
- sliding hip screw is implant of choice for the treatment of both
stable
and
unstable intertrochanteric hip frx;
- may also be used for low
femoral neck frx (basilar neck):
Compression screw fixation for displaced subcapital fracture of the femur.
Success or failure?.
Biomechanical analysis of the sliding characteristics of compression
hip screws.
Treatment of femoral neck fractures with the sliding compression screw.
- Contraindications:
- high
subtroch &
unstable intertroch frx, however, may not be
compressed because of their location in relation to plate, causing
displacement to occur;
- Complications:
- sliding screw fixation of intertroch frxs is assoc w/ incidence of loss
of fixation in < 10% of cases, and most often occurs w/
Unstable frx;
- most fixation failures can be attributed to technical problems of
fracture reduction and screw placement.
- often due to poor reduction, poor implant placement, & inadequate
nail depth;
- use of screw fixation can result in rotation of femoral head on its axis,
leading to further comprimise of the vessels on ligament teres;